signal-to-noise determination

Hi

I want to determinate the LOD of my sample by using the method signal-to-noise ratio 3:1 but I can't find the signal to noise  value on the ChemStation. How can I get this value? Does the software give you also signal-to-noise ratio (2:1 or 3:1)

Thanks in advanced

 

signal to noise

Parents Reply
  • The signal-to-noise calculation would be based off of height, not area. 

     

    You might want to consider using an Intelligent Report. The Performance+Noise.rdl report template will by default report the calculated 6sigma, Peak-to-Peak, and ASTM noise for your defined noise regions as well as calculate Signal/Noise using the 6*SD and USP methods for each peak. 

Children
No Data
Was this helpful?