In the standard report on OpenLAB Chemstation there is a column where i can see if a detected and calibrated peak is in a compound group. The column is named Grp. How can i display compund groups in the Intelligent Report?
Find maintenance, best practice, and troubleshooting articles.
In the standard report on OpenLAB Chemstation there is a column where i can see if a detected and calibrated peak is in a compound group. The column is named Grp. How can i display compund groups in the Intelligent Report?
In Tables->Peaks and Compounds there should be a snippet called "Summed peaks". Drag this into your report. This will show the results for the group compounds.
Hello rhonberg, When I drag this table into my report, there's no result for the concentration ESTD. Only the group name and the aera appears... Could you help me please ?
Hello rhonberg, When I drag this table into my report, there's no result for the concentration ESTD. Only the group name and the aera appears... Could you help me please ?
Hello. Without seeing the method it's hard to say. Have you setup group compounds and assigned a RF or calibrated them?
Hello, thank you for your answer. I solved the problem by adding an other filter "Signal = DAD..." in the table "summed peak"