Automated extraction of EIC

Hi,

I'm using OpenLab CDS 2.2, build 2.20.484.

I'd like to ask about processing LC-MS runs - some of our methods are processed that EIC is extracted automatically when the run/sequence is opened according to compound list in the processing method. However, when I create a new processing method with similar parameters, I'm unable to make the extraction of EICs automated and it requires manually extracting them for each run individually, which becomes inconvenient for longer sequences. I'm unable to find any option in the processing sequence.

 

Also I'd like to ask regarding reporting, if there's any easy way how add on a report a sum of compounds quantified separately? 

 

Thank you.

Parents
  • What would be the best way to send the template? I'd prefer to not disclose it publicly if possible.

  • It is possible to send direct messages to users in the Community if both users are following each other. That might be a good way to coordinate a way to send the template.

     

  • Sounds a good plan. , If this solution works for you, I added you on my following list.

    To keep more confidentiality, ou can also extract only that particular table, as following:

     

    1. Select the table you have, and right-click, then select "Group".

     

     

    2. Once you have a group, around your table (if that was already the case, you can skip the step 1), select this surrounding frame representing the group, and select "Save Composite Group".

     

     

    3. Give a name, and you will see this custom report item in the reporting item list:

     

     

     

     

     

    But also as a file in Content Management under report templates (for Workstation Plus or Client Server systems), or in windows explorer for file based systems:

     

     

    This type of report items are designer tools to help saving custom reporting sections you may want to reuse in other templates.

     

     

    4. You can send me that *.ircg file. 

Reply
  • Sounds a good plan. , If this solution works for you, I added you on my following list.

    To keep more confidentiality, ou can also extract only that particular table, as following:

     

    1. Select the table you have, and right-click, then select "Group".

     

     

    2. Once you have a group, around your table (if that was already the case, you can skip the step 1), select this surrounding frame representing the group, and select "Save Composite Group".

     

     

    3. Give a name, and you will see this custom report item in the reporting item list:

     

     

     

     

     

    But also as a file in Content Management under report templates (for Workstation Plus or Client Server systems), or in windows explorer for file based systems:

     

     

    This type of report items are designer tools to help saving custom reporting sections you may want to reuse in other templates.

     

     

    4. You can send me that *.ircg file. 