Hi,

I'm using OpenLab CDS 2.2, build 2.20.484.

I'd like to ask about processing LC-MS runs - some of our methods are processed that EIC is extracted automatically when the run/sequence is opened according to compound list in the processing method. However, when I create a new processing method with similar parameters, I'm unable to make the extraction of EICs automated and it requires manually extracting them for each run individually, which becomes inconvenient for longer sequences. I'm unable to find any option in the processing sequence.

 

Also I'd like to ask regarding reporting, if there's any easy way how add on a report a sum of compounds quantified separately? 

 

Thank you.

Parents Reply Children
  • Sounds a good plan. , If this solution works for you, I added you on my following list.

    To keep more confidentiality, ou can also extract only that particular table, as following:

     

    1. Select the table you have, and right-click, then select "Group".

     

     

    2. Once you have a group, around your table (if that was already the case, you can skip the step 1), select this surrounding frame representing the group, and select "Save Composite Group".

     

     

    3. Give a name, and you will see this custom report item in the reporting item list:

     

     

     

     

     

    But also as a file in Content Management under report templates (for Workstation Plus or Client Server systems), or in windows explorer for file based systems:

     

     

    This type of report items are designer tools to help saving custom reporting sections you may want to reuse in other templates.

     

     

    4. You can send me that *.ircg file. 

  • It is possible to send direct messages to users in the Community if both users are following each other. That might be a good way to coordinate a way to send the template.

     

  • Hi ,

     

    I modified your report template. 

    In OpenLab CDS v2.2, the "Remove duplicate peak(s)" was not fully compatible with missing compounds or group reporting. We enhanced this from CDS v2.4.

    As this is not out of the box in your CDS v2.2, I have modified your report template to be able to report the group or missing compounds.

     

    Just be aware that with the change I did, you shall not modify the "Remove duplicate peak(s)" as if so, the change I did would be reverted to the default.

    If you need to do this for other tables, you could modify your template just as I did in a text editor, looking for the keyword "distinct". You will see the below expression in your table :

     

     

    Which you should change to the following:

     

     

    I hope this solves your issue.

  • ,

    Good to hear it worked.

    I'm sorry, I forgot to answer this question. When you use an internal table grouping as you do, there would not be a way to display the unit for the summary calculation.

     

    You could add the units for the summary values using a table which is grouped and repeated per compound name. The layout would be slightly different (the summary values will not be part of the table)... An example:

     

  • Thanks a lot, this helped!

    I'm still curious if there's a possibility how to add a unit to the field next to average as indicated on the screenshot:

     