Wiki How To Vent (Shut Down) the GC/MSD - EI - MassHunter
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
Mass Spectrometry requires membership for participation - click to join
  • Wiki
  • Agilent 6100 series LCMS Shutdown Procedure
  • Capabilities of 5977E
  • Collection of GC/MS Resources
  • Collection of LC (HPLC)/MS Resources
  • EI filament P/N G7005-60061 replaces both previous EI filaments
  • Extending Gauge Tube Life
  • Fixing a Corrupt Tune File
  • G6400 Series Shutdown document.
  • GC/MS Part Numbers
  • How to clean an Agilent GCMS Ion Source
  • How to install a capillary column on 5977 Series GC/MSD
  • How To Pump Down (Start) the GC/MSD - EI - MassHunter
  • How To Vent (Shut Down) the GC/MSD - EI - MassHunter
  • Improving GC, GCMS Sensitivity, Dropoff
  • Large 219 peak in autotune
  • Making a robust 7010 method
  • Mass Spectrometry CTA
  • Mass Spectrometry: About
  • Mass Spectrometry: Library Categories
  • Minimum parts needed to add MSD to 6890/7890 (GC Interface parts)
  • Recommended Shut Down Procedure for Agilent TOF/QTOF

How To Vent (Shut Down) the GC/MSD - EI - MassHunter

Agilent Support Insights

 

This Information Applies To: Agilent 5977B GC/MSD with MassHunter GC/MS Acquisition Software

 

Description:

This article describes the procedure to safely vent (shut down) the 5977B GC/MSD with Electron Ionization (EI) 

The MSD will be damaged by incorrect venting. A diffusion pump will push back upstream vaporized pump fluid onto the analyzer if the MSD is vented before the diffusion pump has fully cooled. The turbo pump will be damaged if it is vented while spinning at more than 50% of its normal operating speed.

Steps to follow: 

Agilent recommends that safe temperatures are applied to the GC, GC/MSD interface and MSD to vent the system. If direct communication (DCOMM or LVDS) with the GC is established, you must predefine a Fast Vent method to use this option. If you do not have direct communication (DCOMM or LVDS) between the GC and MSD, create and save a Vent Method with GC oven, Auxiliary heaters (MSD Transfer Line), and inlets turned on and set at room temperature. 

 

  • Remove the analyzer window cover.

  • In the Instrument Control view, Instrument menu, select MS Vacuum Control to display the Vacuum Control dialog.
  • Click Fast Vent if your system supports direct communication between the GC and MSD, or click Vent if it does not, to begin the vent cycle. 

  

 

  • Follow the instructions presented on the screen, and wait until the vent cycle is completed. When the MSD is vented, Do Not put the Agilent MassHunter GC/MS Acquisition software into Instrument Control view. Doing so will turn on the interface heater
  • Close shutoff valves to hydrogen sources that enter the analyzer if the instrument is so equipped.

If hydrogen is used as a carrier gas or JetClean system supply, the carrier and JetClean system supply shutoff valves must be closed before turning off the
MSD power. If the foreline pump is off, hydrogen will accumulate in the MSD and an explosion may occur. Read Agilent GC/MS Hydrogen Safety before
operating the MSD with hydrogen gas.

  • Turn off the MSD by pressing the power switch.
  • Unplug the MSD power chord

Never open the vent valve without first closing the hydrogen supply line to all possible sources of hydrogen entry to the analyzer. This requires closing the hydrogen shutoff valve to the carrier gas flow module and the shutoff valve to the JetClean system supply if either of these hydrogen sources exist on the instrument.

 

  • Turn the vent valve counterclockwise only 3/4 turns or until you hear the hissing sound of air flowing into the analyzer chamber. Do not turn the valve too far, or the O-ring may fall out of its groove.

Ensure the GC oven and the GC/MSD interface are cool before turning off carrier gas flow to prevent damage to the column.

  • You may now turn off the GC and close all the other gas sources.

 

Learn more on how to operate your Agilent 5977B GC/MSD:

Agilent 5977 GC/MSD with MassHunter Workstation e-Learning path on Agilent University 

 

 

Keywords: GCMS, MSD, SQ, 5977B, Turn off GC/MS

 

  • GC-MSD
  • 5977b
  • sq
  • gcms
  • msd
  • agilent support insights
  • Share
  • History
  • More
  • Cancel
Anonymous
Related

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Recommended
Privacy Statement
Terms of Use
Contact Us
Site Help