I have an HP 5973 MSD. When I tried to tune it this morning this error messaged popped up " tune action stopped: gain and offset = 0 and peaks are too narrow. Any suggestions on how to fix it?
Find maintenance, best practice, and troubleshooting articles.
I have an HP 5973 MSD. When I tried to tune it this morning this error messaged popped up " tune action stopped: gain and offset = 0 and peaks are too narrow. Any suggestions on how to fix it?
Hi,
When ion source cleaning performed last time?
Did you change column?
Hi,
When ion source cleaning performed last time?
Did you change column?
The ion source was cleaned in May 2020. The column was changed in January 2020
Check column length at outlet of transferline.
Check how much Volt is showing for EM Horn in your last autotune if u have print out.
How is fliament..
Pump oil level?