Hi,
I have recently purchased the 10ul ultra-microcell Part Number:
Hi,
I have recently purchased the 10ul ultra-microcell Part Number:
Hello:
Remove the cell and run the Instrument Performance tests in the Validate application of the software.
It's possible there are air bubbles in the cell. Reputedly wash wit isopropyl alcohol, followed by your solvent
If they pass, scan at 500 nm or Zero order. The light beam will be visible by placing a a small white piece of paper, (e.g., business card. verify that the beam of light is passing through the aperture of the cell.
Finally, zero the cell on your solvent and scan your sample.
Regards,
Mike
Thank you for your response Mike.
The instrument passes without the cuvette. The cuvette is still experiencing the same issues despite your recommendation. I am unsure if this is relevant, but when I align the sample, the detector only receives approx 18% transmission. The light beam is visible but not as intense and worsens if I try to realign manually.
Best,
Cengiz
Thank you for your response Mike.
The instrument passes without the cuvette. The cuvette is still experiencing the same issues despite your recommendation. I am unsure if this is relevant, but when I align the sample, the detector only receives approx 18% transmission. The light beam is visible but not as intense and worsens if I try to realign manually.
Best,
Cengiz