Hi,
does anybody have experience with analysis of TEOS (Method) and a corresponding (traceable) standard @ Cary630 ?
THX!
Hi,
does anybody have experience with analysis of TEOS (Method) and a corresponding (traceable) standard @ Cary630 ?
THX!
Hi Mhaw,
TEOS (or tetraethyl orthosilicate, Si(OC2H2)4) has many chemical signatures in the mid-infrared spectrum, and the Cary 630 should easily be able to acquire a spectrum of this compound. As TEOS is found in liquid form at room temperature, recommended sampling modules for this compound would be the Dialpath, TumblIR, or if very concentrated, a diamond ATR module.
A recommended traceable standard to use in pharmacopeia for FTIR is NIST SRM 1921b traceable polystyrene film of approximately 35 µm thickness (which is available directly from Agilent under the part number 925-0128).
Please see this Agilent whitepaper for further details: https://www.agilent.com/cs/library/whitepaper/public/whitepaper-pharmacopeia-ftir-5994-2339en-agilent.pdf
Please let me know if you have any further questions!