Anyone knows how to solve this Laser red status?
The Cary 630 doens't allow any analysis with this status!
Thanks
Anyone knows how to solve this Laser red status?
The Cary 630 doens't allow any analysis with this status!
Thanks
Hi Beatriz,
The Laser gain have to be adjusted,
Browse to:
C:\Program Files (x86)\Agilent\MicroLab PC\mTest.exe
and then click on Connect & Scan, raise the gain for the Laser (get around 8500 cts) and then just click on Set to write this to the instrument.
.
Hi Beatriz,
The Laser gain have to be adjusted,
Browse to:
C:\Program Files (x86)\Agilent\MicroLab PC\mTest.exe
and then click on Connect & Scan, raise the gain for the Laser (get around 8500 cts) and then just click on Set to write this to the instrument.
.
Thanks
I will do it
It works
Thanks for the fast and clear support.