Cary 630 Red or Low Energy

Good Morning,

Would you please advise what are the possible causes of getting a low energy value of 16 on my cary 630?

Parents
  • Hi chandrasekhar,

    Please forgive me if you were asking for the current gain value in this case and not what the gain values is.  The Cary 630 amplifies the signal from the detector. If you are familiar with Photomultiplier Tubes, this is a similar concept. The application of the signal is represented by the gain value. Increasing it will increase the energy.

    What gain value is good depends upon your accessory. When you put an accessory on the Cary 630 for the first time, the software will automatically optimize the gain value for the accessory. Sometimes you need to make an adjustment as the amount of energy getting to the detector changes. The default value can also be overridden in the method settings.

    Please let me know if you have any questions or concerns.

    Best Regards,

    Andrew Coukouma

Reply
  • Hi chandrasekhar,

    Please forgive me if you were asking for the current gain value in this case and not what the gain values is.  The Cary 630 amplifies the signal from the detector. If you are familiar with Photomultiplier Tubes, this is a similar concept. The application of the signal is represented by the gain value. Increasing it will increase the energy.

    What gain value is good depends upon your accessory. When you put an accessory on the Cary 630 for the first time, the software will automatically optimize the gain value for the accessory. Sometimes you need to make an adjustment as the amount of energy getting to the detector changes. The default value can also be overridden in the method settings.

    Please let me know if you have any questions or concerns.

    Best Regards,

    Andrew Coukouma