i have encountered an error with Cary 100 UV-VIS, E9328 fatal initialisation, fault reported by the instrument software while calibrating, error code:9:9328SEQ:E9328 fatal initialisation, this error occurs when initialising wavelength
i have encountered an error with Cary 100 UV-VIS, E9328 fatal initialisation, fault reported by the instrument software while calibrating, error code:9:9328SEQ:E9328 fatal initialisation, this error occurs when initialising wavelength
Hello
A lot of things are related to this error message. I'd like to recommend to check from start point(Lamp)
1. Lamp
2. Solenoid valve for changing lamp
3. Zero order alignment(set the dial "15", and rotate the micrometer by hand to move it to zero order position)
4. Micrometer
B.Regards
CSH