The Cary 845x UV-Visible system software does not have a way to export "All Spectrum". How can I export all of the spectrum scans that were collected during a kinetic run. There are hundreds. This is way too many to export one by one.
The Cary 845x UV-Visible system software does not have a way to export "All Spectrum". How can I export all of the spectrum scans that were collected during a kinetic run. There are hundreds. This is way too many to export one by one.
The customer has resolved his problem with a macro form the UCL folder on the ChemStation disk. (These are non-supported macros, but were likely tested at the time of contribution.)
In this case , Macro Contribution #4
UV ChemStations Revision A.05.xx
Date 08/05/98
Mode : KINETICS
Contribution 'Macro functions are provided to export kinetics
data from one cell to a text file. A dialog is
offered to reduce the data set by specifying a
subset of the acquired wavelengths and the
acquisition times.'
Well done Michael4008!
All previous replies were inadvertently deleted.
Regards,
Mike