i have encountered an error with Cary 100 UV-VIS, E9328 fatal initialisation, fault reported by the instrument software while calibrating, error code:9:9328SEQ:E9328 fatal initialisation, this error occurs when initialising wavelength
i have encountered an error with Cary 100 UV-VIS, E9328 fatal initialisation, fault reported by the instrument software while calibrating, error code:9:9328SEQ:E9328 fatal initialisation, this error occurs when initialising wavelength
Hello tata,
It could also due to a bad signal from the lamp (lamps need to be replaced or aligned). You can check the %T in Align application just to know if the detector can see something.
Best regards,
Rémi
Hello tata,
It could also due to a bad signal from the lamp (lamps need to be replaced or aligned). You can check the %T in Align application just to know if the detector can see something.
Best regards,
Rémi