Hello Agilent Community,

We recently acquired the Cary 630 with the ATR, but I'm trying to access the MicroLab PC and this message appears:

"object reference not set to an instance of an object"

The equipament was installed and qualifyied a couple weeks ago, but now none of the user can acess the software.

The MicroLab Quant starts without any problems.

What could be causing this error?

Parents
  • Hi..

    You need to check in result folder under report section , whether the data path saved  directly into root directory of the database, if so then make one sub folder in database & add into path. 

    Also, just ensure proper report template selection as per method type. 

    In SCM Group, Maximum 3 Databases are allowed so add in following Priority PublicDB>SystemCheckDB>Local DB. 

    This should work. 

Reply
  • Hi..

    You need to check in result folder under report section , whether the data path saved  directly into root directory of the database, if so then make one sub folder in database & add into path. 

    Also, just ensure proper report template selection as per method type. 

    In SCM Group, Maximum 3 Databases are allowed so add in following Priority PublicDB>SystemCheckDB>Local DB. 

    This should work. 