How to solve step issue for reflectance %R

Hi All,

 

I try to run program wavelength ( 500 - 2500%R ), But during the proses there was a "STEP" come out like as per attachment,

i keep playing with "Detector Changeover&Gratis Changeover" but still cannot solve, normal running also happen like this@@

 

Please Help!!!

 

 

Thanks

ZulGFTECH

Parents
  • One of the ways to eliminate the "step" at the grating changeover is to use the Baseline Correction feature. In the method Setup, in the Baseline Tab.

    Select Baseline. Set the other parameters as per your method. Clock OK to return to the main Scan Window.

    From the main Scan window, the Baseline button will now be active. Press the Baseline button and run the 100%R Baseline using your blank.

    Then run your sample as normal.

Reply
  • One of the ways to eliminate the "step" at the grating changeover is to use the Baseline Correction feature. In the method Setup, in the Baseline Tab.

    Select Baseline. Set the other parameters as per your method. Clock OK to return to the main Scan Window.

    From the main Scan window, the Baseline button will now be active. Press the Baseline button and run the 100%R Baseline using your blank.

    Then run your sample as normal.