Using a 8453 UV-Vis in standard mode on Chem Station, do you know if absorbance blank is subtracted from sample when result is displayed after reading?
Using a 8453 UV-Vis in standard mode on Chem Station, do you know if absorbance blank is subtracted from sample when result is displayed after reading?
Hello,
I just wanted to follow-up with you to see if Mike's reply answered your question. If so, it would be appreciated if you could be return to this thread and mark a Correct Answer.
Yes. Thank you!
Aurora
[Personal Information Removed by Moderator]
Thank you for the update! I will go ahead and mark Mike's reply as the Correct Answer.
Thank you for the update! I will go ahead and mark Mike's reply as the Correct Answer.