QTof, Qual exclude MS/MS experiments with low abundance

Hi,

I am using Qual 12 for LC-QTof. If I use Find-by-Formula I receiced hits with multiple MS/MS-experiments. In this case 11 scans.

Where can I set the parameters that I only have the 4 largest experiments (green) and exclude the red one. That's important for the average MS/MS spectrum.

Thanks!

Parents
  • Hello  ,

    From what I can find, only the MS spectrum can be limited to the top x percent of the peak. This setting is on the Find by Formula->Peak Spectrum (MS) tab and is only available if you have enabled Show Advanced Settings in the Configuration menu 

    I'm not finding a way to limit the MS/MS spectrum averaging directly, but you can set thresholds on the Fragment Peak Filter tab that can clean up the averaged MS/MS spectrum.

    Here is an example from the demo data showing the MS/MS spectrum of a peak with no thresholds set.

    And here is the same spectrum using the default threshold settings.

    For this data it is not a very dramatic difference, but it should be possible to adjust these values to help clean up your spectrum depending on your data.

Reply
  • Hello  ,

    From what I can find, only the MS spectrum can be limited to the top x percent of the peak. This setting is on the Find by Formula->Peak Spectrum (MS) tab and is only available if you have enabled Show Advanced Settings in the Configuration menu 

    I'm not finding a way to limit the MS/MS spectrum averaging directly, but you can set thresholds on the Fragment Peak Filter tab that can clean up the averaged MS/MS spectrum.

    Here is an example from the demo data showing the MS/MS spectrum of a peak with no thresholds set.

    And here is the same spectrum using the default threshold settings.

    For this data it is not a very dramatic difference, but it should be possible to adjust these values to help clean up your spectrum depending on your data.