Sample quantitation failed

Hello everyone

sometimes when I create a new result table in mass hunter quantitative 8 this message appear "Sample quantitation failed. Not ready for quantitation. Please calibrate before quantitating."

The only thing I do is create a new result table for a new calibration and call the method from the previous result table. This doesn't happen all the time but occasionally

Thanks

Parents
  • Same problem here. I followed this procedure step by step but could not make it work.


    1. Changed my samples to Cal and levelled them.

     


    2. Created levels from Cal.Samples and added the concentrations.




    3. Saved> Analyzed > Edit Method > Saved again. I do see the calibration curve, but...



    4. If I "Quantitate Batch," I am getting the same Error



    MassHunter Quantitative Analysis 10.2
    Tried it with and without samples. With and without averaging the calibrations.
    I checked multiple explanatory manuals about this and have no clue why it won't work.

    I hope I am not missing something and would greatly appreciate any support!

  • The only way I solved this problem, is sending again the calibration. It seems there is something wrong in the original batch on the instrument.