MassHunter shows background spectrum as compound spectrum

I don't know why in Compound detail view I see background spectrum as my compound spectrum every other time I click on a hit. I don't know if this is a software bug or there is option to remove it. But it exactly happens every other time I click on a spectrum hit. It doesn't affect anything, but it's annoying because I need to click on the same hit three times to show the correct spectra of my compound. Any help is greatly appreciated.

Thanks,

Ehsan

Parents Reply
  • Hi  ,

    We were able to duplicate what you are seeing by switching to another compound hit. The average scan spectrum shows the first time a different hit is selected, the second click clears the spectra, and the compound spectra is displayed on the third click.  However, If you click on the title of the average scan spectrum (top pane) in the Spectral Difference Results, the Compound Spectrum will load immediately. 

    I hope this helps as there are no plans to change this behavior in B.07 or B.08.  This behavior doesn't not happen in MassHunter 10.