Hello,
I was wondering if you know a template that would only generate the report.results.xml ?
If not, could I write or modify a template to do so ?
Thank you for your help !
Find maintenance, best practice, and troubleshooting articles.
Hello,
I was wondering if you know a template that would only generate the report.results.xml ?
If not, could I write or modify a template to do so ?
Thank you for your help !
Hello,
What software are you asking about Chemstation, EZChrome, or CDS 2.x?
Marty Adams
Data analysis and reports are generated using Quant my way (LCMS) and data acquisition is done through MassHunter 10.1 Build 10.1.733.0
Hi Canadian_Chemist ,
The report template does not create the report.results.xml file. That is created by quant. You can specify it always be created in the reporting method.
You would still need to specify a report template in your method. I believe you could essentially make an empty report builder template and the results file should still be created.
It may also be possible to write a python based report that would duplicate the report.results.xml or at least what you needed. You could look at the Gen_LIMs_xmlOutput.repot.xml template for an example.
Hi Canadian_Chemist ,
The report template does not create the report.results.xml file. That is created by quant. You can specify it always be created in the reporting method.
You would still need to specify a report template in your method. I believe you could essentially make an empty report builder template and the results file should still be created.
It may also be possible to write a python based report that would duplicate the report.results.xml or at least what you needed. You could look at the Gen_LIMs_xmlOutput.repot.xml template for an example.