Index was out of range. Must be non-negative and less that the size of the collection
Find maintenance, best practice, and troubleshooting articles.
Index was out of range. Must be non-negative and less that the size of the collection
What version of GCMS MassHunter Acquisition are you using?
What version of GCMS MassHunter Acquisition are you using?
version B07.05
If you will give your local Agilent Software Support office a call, they will most likely need to see sequence_info.log If running in a sequence. This is located in \masshunter\gcms\x\. X represents instrument 1,2,3,4 that is configured.
If not running a sequence and just loading the method, does acquisition close by itself?
method loading is no problem, error appears when downloading method on MS
Then the Local Support office will need to run a log file to see what is going on. Do all methods have the issue?