MassHunter Workflow B.08.00; Compound list contains compounds but does not show them. See picture.

We are unable to add components to the compound List, we can select the correct one but they are not shown. Can anyone assist us in this problem?

Thanks in advance!

Parents
  • Hello ,

    It appears that the compound list is being filtered on a column that has now been hidden. It is indicated in the window title and tab, but it is not easily seen.

    Qual does not reset filtering even if you load another method or close and restart qual. You can either add back the m/z column and remove the filter, or you can right click on one of the column headers and choose Restore Default Table Layout to quickly remove all filters.

Reply
  • Hello ,

    It appears that the compound list is being filtered on a column that has now been hidden. It is indicated in the window title and tab, but it is not easily seen.

    Qual does not reset filtering even if you load another method or close and restart qual. You can either add back the m/z column and remove the filter, or you can right click on one of the column headers and choose Restore Default Table Layout to quickly remove all filters.