Can someone walk me through how to set up a calibration curve in Chemstation version: MSD Chemstation E.02.01? The GC I am using is a GC-ECD 6890 Series.
The attached document is how to setup a quantitation database for gc signals. In the document ignore all instances of 2nd signal. If still have questions you give your local Agilent Software Support office a call.
The attached document is how to setup a quantitation database for gc signals. In the document ignore all instances of 2nd signal. If still have questions you give your local Agilent Software Support office a call.
Hi,
This is actually perfect because the method i am doing is a dual column. I am having a bit of an issue though. I do the first cal point and that goes perfectly, but when i do the rest of the cal points it doesn't go so smoothly. I follow the directions for the second, third and fourth, etc points and instead of updating the level it duplicates and triplicates the compounds in database. Also when i go to quantitate the rest of the cal points it doesnt pick up the second column just shows me the results for the first column. Am i missing something? I have been working on this all day and i cant seem to see what i am missing.
I would advise you give the local Agilent Software Support Center a call.
i was able to figure this out, i misread a section of the directions. Thank you for your help.