I have Report Builder 11.0. Now I can't open my old templates. I get the following error:
'Cannot activate designer
'Agilent.MassSpectrometry.DataAnalysis.CustomWorkflowReport.DataSourceDesigner_10_0, RptBldReportEngine'
How can I fix this?
I have Report Builder 11.0. Now I can't open my old templates. I get the following error:
'Cannot activate designer
'Agilent.MassSpectrometry.DataAnalysis.CustomWorkflowReport.DataSourceDesigner_10_0, RptBldReportEngine'
How can I fix this?
Hello Imma Ferrer,
What type of template are you attempting to open? If this is a Qual template RB 11 would not work with those. For Qual templates you would use RB 10.1, which should still be installed.
It is a Qual template apparently created with version 10.1. But I realize I do not have this version anymore, I only have 10.2, 10.2.1 and 11.0 for some reason.
See hexatriene's reply in the other thread for how to get RB 10.1 back.
(+) Why are my pdf Qual report templates no longer readable with report builder? - Forum - Mass Spectrometry Software - Agilent Community
Also that discussion has a clearer explanation of why you need a specific revision to open a template. The version must match what is listed as the <EditorVersion> section in the template, which can be seen by opening the template in a simple text editor.
Thanks for all your help! I was able to re-install version 10.1 and I can open the templates again.