Intelligent Reflex Setup

I am trying to set up intelligent reflex where if an analyte is >Calibration Level 2, the sample will reflex to a separate method.  (I think) I was able to make the necessary changes under the DA tab in the acquisition method. Can anyone point me in the right direction on what I need to set up in the quant method to make this possible?

  • Hello  ,

    If you are using the Above Cal. Range reflex type then it is not possible to specify a different acq method. You can only specify a redo injection volume. This will only be triggered if the sample measures above the calibration range of your specified Quant method. You can't, for example, have a 3 level calibration method and then specify to re-inject if the response is above Level 2. You would need separate Quant method were Level 2 is the highest level and specify this for your Intelligent Reflex Quant Method.