I have a question about Ultivo QQQ , what is the accuracy of this instrument? Can I get expected mass ± 0.2 m/z in each analysis?
Find maintenance, best practice, and troubleshooting articles.
I have a question about Ultivo QQQ , what is the accuracy of this instrument? Can I get expected mass ± 0.2 m/z in each analysis?
Hi Fereshteh ,
The specifications document of the G6465B Ultivo shows a Mass accuracy of:
0.1 Da from m/z 5 to 1,000
0.2 Da from m/z 1,000 to 1,400
The Mass resolution full width at half maximum (FWHM) is 0.7 Da.
I hope this answers your question.
Regards,
Hi Philippe,
Thanks for the response. Yes, I was looking for specification of the instrument and could not find it. Any idea where can I find an official document from Agilent about accuracy and resolution?
Thanks,
Fereshteh