The capillary voltage on our G6420A QQQ is set for 3000V but the actual is around 40V and I am not getting peaks. What is most likely wrong with the instrument? Thanks
Find maintenance, best practice, and troubleshooting articles.
The capillary voltage on our G6420A QQQ is set for 3000V but the actual is around 40V and I am not getting peaks. What is most likely wrong with the instrument? Thanks
There could be a problem with the platinum end of the capillary or one of the canted coils springs is not making contact. Check the connections on the 2 wires at the top of the spray shield. If there is a connection there measure at that point. center wire should be 3K and the left wire or spay shield should be 2500
There could be a problem with the platinum end of the capillary or one of the canted coils springs is not making contact. Check the connections on the 2 wires at the top of the spray shield. If there is a connection there measure at that point. center wire should be 3K and the left wire or spay shield should be 2500