System Tune LC/QTOF 6545XT

Hello everyone,

I am facing an issue when trying to perform a System Tune in negative mode of both Quadrupole and TOF following 1 month of the QTOF being off due to a hard drive failure.

Has anyone encountered that error and was able to fix it ?

Backup current tune file before autotune
Running QTOF System Tune in Negative Polarity
Begin Running System Tune
Backup All Parameters
Spraying Calibrant Solution
Disable sureMass optimization
Setup Dual AJS ESI Source to Achieve Stable Signals for Auto Tune
Reset All Parameters to Their Default Values
      Coarse Estimate Low Gain Amp Offset Found at 34153
         Final Optimum Low Gain Amp Offset Found at 34170
      Coarse Estimate Hi Gain Amp Offset Found at 31228
         Final Optimum Hi Gain Amp Offset Found at 31174
   Waiting for Calibrant to Elute...
   Monitoring 601.81 m/z Peak for Intensity and Stability
      Signal Intensity and Stability Satisfied Auto Tune Requirements
   Performing Coarse TOF Mass Calibration Using 6/10 Calibrant Masses 
      Successfully Perform Mass Calibration with RMS Deviation (ppm): 0.82
   Reset Optics 2 and TOF Parameters to Their Default Values
      Signal of the Highest m/z Peaks Below Requirement, Increase MCP to 900
   Coarse TOF Resolution - Mirror Mid Optimization
      Change Mirror Mid from 1,722.0 to Optimum Value at 1,717.2
   Coarse TOF Resolution - Puller Offset Optimization
      Change Puller Offset from -32.0 to Optimum Value at -36.9
   Performing Coarse TOF Mass Calibration Using 6/10 Calibrant Masses 
      Successfully Perform Mass Calibration with RMS Deviation (ppm): 0.02
   Signal below target do not need to adjust VCap 4000 observed peak height 2296
   Adjusting Quad AMU to Produce Single Ion Event at MCP of 720
      Attenuate the signal to the highest Quad AMU setting, probably the MCP value is too low
      Increasing MCP value by 50 and try again
   Signal below target do not need to adjust VCap 4000 observed peak height 18663
   Adjusting Quad AMU to Produce Single Ion Event at MCP of 770
      Ion arrival frequency 29.9 % too high for the peak height 4483
      Increasing MCP value by 25 and try again
   Signal below target do not need to adjust VCap 4000 observed peak height 50977
   Adjusting Quad AMU to Produce Single Ion Event at MCP of 795
      Ion arrival frequency 24.1 % too high for the peak height 4495
      Increasing MCP value by 25 and try again
   Adjusting VCap to reduce signals to reasonable height 100000
   Reduce high mass signals by reduce VCap to 2875 V
   Adjusting Quad AMU to Produce Single Ion Event at MCP of 820
      Ion arrival frequency 18.1 % too high for the peak height 4072
      Increasing MCP value by 25 and try again
   Adjusting VCap to reduce signals to reasonable height 100000
   Reduce high mass signals by reduce VCap to 2219 V
   Adjusting Quad AMU to Produce Single Ion Event at MCP of 845
      Coarse optimum Quad AMU Found at 11.6 with Peak Height of 4151
      Optimum Quad AMU Found at 1.0 with Arrival Frequency of 4.6 %
   Searching MCP to Produce 20 Counts per Single Ion
           Counts         MCP     Frequency               Status
           04.7              845        7.7 %                in Progress
           09.7              891        9.6 %                in Progress
           13.8              922        9.6 %                in Progress
           24.0              940        8.7 %                in Linear Range
Previous Estimation was NOT correct, Readjust Single Ion Event
   Adjusting VCap to reduce signals to reasonable height 100000
   Reduce high mass signals by reduce VCap to 1750 V
   Adjusting Quad AMU to Produce Single Ion Event at MCP of 936
      Coarse optimum Quad AMU Found at 11.1 with Peak Height of 4629
      Optimum Quad AMU Found at 3.9 with Arrival Frequency of 3.9 %
   Searching MCP to Produce 20 Counts per Single Ion
           Counts         MCP     Frequency               Status
           08.5              936        8.1 %                in Progress
           09.8              971        7.8 %                in Progress
   Trying to set MCP to 1001.79694352159, which is out of bounds, set to maximum: 1000
           11.6              1000        9.4 %                in Progress
   Trying to set MCP to 1025.27, which is out of bounds, set to maximum: 1000
           13.4              1000        7.9 %                in Progress
   Trying to set MCP to 1019.86, which is out of bounds, set to maximum: 1000
           14.3              1000        8.3 %                in Progress
   Trying to set MCP to 1017.11, which is out of bounds, set to maximum: 1000
           10.1              1000        9.9 %                in Progress
   Trying to set MCP to 1029.79069767442, which is out of bounds, set to maximum: 1000
           12.1              1000        8.0 %                in Progress
   Trying to set MCP to 1023.82, which is out of bounds, set to maximum: 1000
           10.4              1000        8.2 %                in Progress
   Trying to set MCP to 1028.8438538206, which is out of bounds, set to maximum: 1000
           16.1              1000        7.9 %                in Linear Range
   Trying to set MCP to 1003.93, which is out of bounds, set to maximum: 1000
           11.1              1000        9.3 %                in Progress
   Trying to set MCP to 1026.7, which is out of bounds, set to maximum: 1000
           09.6              1000        7.7 %                in Progress
   Trying to set MCP to 1031.33, which is out of bounds, set to maximum: 1000
           15.7              1000        7.7 %                in Progress
   Trying to set MCP to 1012.82, which is out of bounds, set to maximum: 1000
           10.9              1000        8.4 %                in Progress
   Trying to set MCP to 1027.33, which is out of bounds, set to maximum: 1000
           13.7              1000        9.0 %                in Progress
   Trying to set MCP to 1018.81727574751, which is out of bounds, set to maximum: 1000
           11.9              1000        8.8 %                in Progress
   Trying to set MCP to 1024.40728476821, which is out of bounds, set to maximum: 1000
           12.3              1000        8.7 %                in Progress
   Trying to set MCP to 1023.16, which is out of bounds, set to maximum: 1000
           12.4              1000        8.3 %                in Progress
   Trying to set MCP to 1022.79, which is out of bounds, set to maximum: 1000
           16.0              1000        9.6 %                in Linear Range
   Trying to set MCP to 1003.99, which is out of bounds, set to maximum: 1000
           15.7              1000        8.7 %                in Progress
   Trying to set MCP to 1012.9, which is out of bounds, set to maximum: 1000
           10.3              1000        7.5 %                in Progress
   Restore to previous MCP  value 720
Disable sureMass optimization
Stop Spraying Calibrant Solution
System Tune did not complete due to the following reason(s): 

Cannot Find the Optimum Detector Gain
Completed QTOF System Tune in Negative Polarity

Many thanks

  • Hi,

    As per the autotune record, the detector MCP is maxing out at 1000v. It cannot get enough signal to achieve the detector calibration.  You can see that from the beginning at line #19 where detector MCP jump from previous 720v to 900v.

    Line #19  Signal of the Highest m/z Peaks Below Requirement, Increase MCP to 900

     Line #60 Searching MCP to Produce 20 Counts per Single Ion

               Counts         MCP     Frequency               Status

               08.5              936        8.1 %                in Progress

               09.8              971        7.8 %                in Progress

      Line #64  Trying to set MCP to 1001.79694352159, which is out of bounds, set to maximum: 1000

               11.6              1000        9.4 %                in Progress

     

    Check the basic first, enough tune mix solution in bottle, inspect nebulizer needle for physical damage and proper adjustment.  Check the tune mix flow out of the peek tubing on top of nebulizer, disconnect the tubing, with bottle B to On, you should have 1 drop falling off every 7-10sec.

    Check the result in positive mode from MCP value.  If it’s close to max 900-1000v, like negative mode that may indicated the detector end of life, may need replacement.  But If your Positive mode passes with normal MCP value750-850V, but the Negative mode failed, that may indicate issue with negative mode ion generation at the front end of the QTOF.  I suggest you place a service call at Agilent in both cases.

    Best regards

  • Thanks for your help.

    There seems indeed to be an issue linked to the detector: after testing it in Positive mode, the tuning doesn't pass and MCP is maxing out as well.
    In any case, I will contact Agilent.

    Best regards