Wiki Forced Cold Start Procedure
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
Liquid Chromatography requires membership for participation - click to join
  • Wiki
  • About LC and CE
  • Backflushing the needle seat of the Agilent Vialsampler
  • Best Practices for Using an Agilent LC System
  • Can the 1120 and 1220 Series LC Systems be Controlled by the Instant Pilot?
  • Collection of LC (HPLC) Resources
  • Distance to window on CE capillary
  • Forced Cold Start Procedure
  • Fraction Collector Delay Volume
  • General Column Regeneration Procedure
  • How to Add a New System in Lab Advisor
  • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
  • How to change the needle and needle seat in an Agilent G7129x Vialsampler
  • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
  • How to Install Lab Advisor
  • How to perform a cold start for 1200 series LC modules
  • How to Prepare your HPLC for a Long Time Shut Down
  • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
  • How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module
  • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
  • How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
  • How to replace the PTFE frit of the Purge Valve in LC Pumps
  • How To Set Screen Brightness on the Instant Pilot Controller
  • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pumps
  • InfinityLab Quick Connect A-Line Fittings
  • Join the Agilent Biopharma Influencers Community!
  • LC Error Codes
  • LED Status Indicators
  • Method Scouting Wizard Tutorial Videos now available on YouTube!
  • Needle seat backflush procedure
  • Normal Phase seals and wear-in procedure
  • Optical Balance Procedure for Refractive Index Detector
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
  • Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
  • Spectra are not showing for some peaks when collected only with Peak Apex
  • Troubleshooting High Pressure Issues in Agilent LC Systems
  • Troubleshooting peak area variation between injections
  • Use of Multiple Degasser Chambers for Isocratic and Binary pumps without SSV

Forced Cold Start Procedure

Agilent Support Insights

 

This Information Applies To: All Agilent HPLC modules (except hosted modules).

 

Issue:

If a module does not initialize correctly or it is not recognized by the software it is recommended to call our Remote Support Engineers to troubleshoot the issue. You may be asked to follow the cold start procedure below to recover the correct functionality of the module. A forced cold start can be used to bring the module into a defined mode with default parameter settings.

 

Before starting:

  • Save your methods and data before executing a forced cold start. A forced cold start erases all methods and data stored in the non-volatile memory of the module, i.e a non saved state.
    Exceptions are calibration settings, diagnosis and repair log books which will not be erased.

 

Do not use a pencil or a fragile tool to move the switches to avoid breaking it inside the module. We recommend you to use a screwdriver or a pen.

 

Steps to follow:

  • Switch OFF the modules.
  • Disconnect the cables that connect the troubled module to the others.
  • Take note of the current configuration of the switches which are on the back of the module inside the CONFIG slot (Figure 1). This is important, because after the procedure you will have to move the switches back to that original configuration. Your module might have 2, 6 or 8 switches depending on the model. The 6-switch configuration is for all Infinity II Series modules except the Hosted modules that use a 2-switch configuration (Multicolumn Thermostat, Flexible Cube, UIB II, and 1290 Infinity Valve Drive), while the 8-switch configuration is for 1100 and 1200 Series instruments.

 

Figure 1. Location of Configuration Switch (CONFIG) and different type of Switches of HPLC modules.

 

  • Change the switches configuration according to Table 1 below.

 

2 switches 0 1
6 switches 1 0 0 0 0 1
8 switches 1 1 0 0 0 0 0 1

Table 1: Switches' configuration for a forced cold start. 1 means up, 0 means down.

 

 

  • Switch ON the instrument and wait 3 minutes until the initialization is completed.
  • Switch OFF the instrument.
  • Place the switches in their original position.
  • Reconnect all the cables and switch ON all the modules in the stack.

 

Special considerations: 

The instrument will always revert to the default state after a cold start. This is especially important for plate types and the metering head size in autosamplers. For example, the 900 µL analytical head upgrade will not be recognized. It must be set again in the software (the same as during initial installation). The sampler will default to the standard 100 µL head.

 

Learn more on how to operate your Agilent LC System:

LC User manual library on Agilent.com

 

Keywords: stuck autosampler, LC error, memory, connection, non-volatile memory, board LAN, boot main, default settings

 

 

  • methods
  • LC
  • etc.
  • Detectors
  • Fraction Collectors
  • scripts
  • Pumps
  • lc 1290 infinity ii
  • 1260 infinity
  • 1290
  • 1120
  • sample introduction
  • agilent support insights
  • 1220 infinity ii
  • 1200
  • templates
  • 1100
  • 1290 infinity ii
  • Vacuum Degassers