Wiki LED Status Indicators
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
Liquid Chromatography requires membership for participation - click to join
  • Wiki
  • About LC and CE
  • Backflushing the needle seat of the Agilent Vialsampler
  • Best Practices for Using an Agilent LC System
  • Can the 1120 and 1220 Series LC Systems be Controlled by the Instant Pilot?
  • Collection of LC (HPLC) Resources
  • Distance to window on CE capillary
  • Forced Cold Start Procedure
  • Fraction Collector Delay Volume
  • General Column Regeneration Procedure
  • How to Add a New System in Lab Advisor
  • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
  • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
  • How to Install Lab Advisor
  • How to perform a cold start for 1200 series LC modules
  • How to Prepare your HPLC for a Long Time Shut Down
  • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
  • How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module
  • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
  • How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
  • How to replace the PTFE frit of the Purge Valve in LC Pumps
  • How To Set Screen Brightness on the Instant Pilot Controller
  • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pumps
  • InfinityLab Quick Connect A-Line Fittings
  • Join the Agilent Biopharma Influencers Community!
  • LC Error Codes
  • LED Status Indicators
  • Method Scouting Wizard Tutorial Videos now available on YouTube!
  • Needle seat backflush procedure
  • Normal Phase seals and wear-in procedure
  • Optical Balance Procedure for Refractive Index Detector
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
  • Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
  • Spectra are not showing for some peaks when collected only with Peak Apex
  • Troubleshooting High Pressure Issues in Agilent LC Systems
  • Troubleshooting peak area variation between injections
  • Use of Multiple Degasser Chambers for Isocratic and Binary pumps without SSV

LED Status Indicators

Agilent Support Insights

 

This Information Applies To: All HPLC modules

 

Overview:

The LED Status Indicators are common features for all the HPLC modules. The indicators are located on the top right corner of each module, opposite to the power switch led as shown in Figure 1.

 

They can vary between six different statuses: Grey (led off), Green, Yellow, Red, flashing Yellow or flashing Red.

 

The LED Status is very informative on the module performance and can provide important information for troubleshooting.

 

 

 

Figure 1: Position of LED status indicators and the Power Switch on HPLC modules.

 

 

Resolution: 

  • Red LED

Error condition.

It indicates that the module has detected an internal problem that will effect the correct operation of the module. Usually, an error condition requires attention (e.g. leak, defective internal components). An error condition always interrupts the analysis. If the error occurs during analysis, it is propagated within the LC system, i.e. a red LED may indicate a problem of a different module. Use the status display of your user interface for finding the root cause/module of the error.

 

An error code can either pop-up in your CDS software, or can be found in Lab Advisor under the Log & Results panel (as shown in figure 2).

Lab Advisor will also provide information on the error source and suggest solutions.

 

 

 

 

Figure 2: Error cause identification using Lab Advisor.

 

  • Yellow LED

Not-ready condition.

The module is waiting for a specific condition to be reached or completed (for example, immediately after changing a setpoint), or while a self-test procedure is running.

The LED is yellow while the module is initializing, after the initialization process is completed the led normally turns off/grey.

If the yellow color remains for longer than few minutes, it may indicate that the module has an issue trying to reach one of the method setpoints.

 

  • Green LED

Run mode. A green status indicator, indicates the that module is performing an analysis.

 

  • Grey LED (led off)

Idle status. This status does not indicate a failure or an error.

When the status indicator is OFF (and Power Switch light is on), the module is in a pre-run condition, and it is ready to begin an analysis.

 

  • Blinking yellow/red

Resident firmware mode or low lever error mode.

A blinking LED indicates that the module is in resident mode (e.g. during update of main firmware).


A fast blinking LED indicates that the module is in a low-level error mode. In such a case, try to reboot the module or try a cold-start (Forced Cold Start: to solve module initialization issues and software recognition). Then try a firmware update according to the User Manual. If a firmware update does not help, a main board replacement may be required: contact Agilent your local Agilent Service representative for further support.

 

Learn more on how to operate your Agilent LC System:

LC User manual library on Agilent.com

 

 

Keywords: LED, status indicators, error, idle, green, yellow, red, blinking

 

 

  • liquid chromtography
  • troubleshooting
  • HPLC
  • led indicator
  • led status
  • Share
  • History
  • More
  • Cancel