  • Backflushing the needle seat of the Agilent Vialsampler
  • Best Practices for Using an Agilent LC System
  • Can the 1120 and 1220 Series LC Systems be Controlled by the Instant Pilot?
  • Collection of LC (HPLC) Resources
  • Distance to window on CE capillary
  • Forced Cold Start Procedure
  • Fraction Collector Delay Volume
  • General Column Regeneration Procedure
  • How to Add a New System in Lab Advisor
  • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
  • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
  • How to Install Lab Advisor
  • How to perform a cold start for 1200 series LC modules
  • How to Prepare your HPLC for a Long Time Shut Down
  • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
  • How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module
  • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
  • How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
  • How to replace the PTFE frit of the Purge Valve in LC Pumps
  • How To Set Screen Brightness on the Instant Pilot Controller
  • InfinityLab Quick Connect A-Line Fittings
  • Join the Agilent Biopharma Influencers Community!
  • LC Error Codes
  • LED Status Indicators
  • Method Scouting Wizard Tutorial Videos now available on YouTube!
  • Needle seat backflush procedure
  • Normal Phase seals and wear-in procedure
  • Optical Balance Procedure for Refractive Index Detector
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
  • Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
  • Spectra are not showing for some peaks when collected only with Peak Apex
  • Troubleshooting High Pressure Issues in Agilent LC Systems
  • Troubleshooting peak area variation between injections
  • Use of Multiple Degasser Chambers for Isocratic and Binary pumps without SSV

Backflushing the needle seat of the Agilent Vialsampler

Agilent Support Insights

 

This Information Applies To: Agilent Infinity II Vialsamplers.

 

Issue:

High pressure in the Vialsampler is oftentimes due to a blockage in the needle seat assembly. Sample particulates or plugs from the vial caps can be picked up by the needle when the Vialsampler is drawing the sample. When the valves switches back to mainpass, they are forced into the seat capillary and cause obstruction and high pressure.

 

Steps to follow: 

To backflush the needle seat:

  1. Turn on the pump flow rate to 3 mL/min through your CDS Software, Lab Advisor or through the Instant Pilot and open the purge valve. Set the injector to bypass mode.
  2. Using Lab Advisor (downloadable by copy-pasting http://labadvisor.pl29.agilent.com/latest to a new browser tab or window), go to the "Service and Diagnostics" section, highlight the Vialsampler, and choose "Injector Steps". Next, click "Needle up" for the Vialsampler to lift the needle (see Picture 1) up fully.
    Picture 1: Lab Advisor Injector Steps - Needle Up.
  3. Remove the safety cover (see Picture 2), then lift up and flip the wash port to the left into the service position (Picture 3).
    Picture 2: Removal of the safety cover.

    Picture 3: Moving the wash port into the service postion.

    Safety Warnings: Avoid removing capillaries under high pressure, make sure to wear appropriate protective gear.

  4. Disconnect capillary in port 6 of the injection valve. Disconnect the seat capillary from port 5 in the injection valve and connect the needle seat capillary to port 6, so that the flow will go directly from the pump through the needle seat (make sure to work in bypass mode, use Picture 4 as a reference).

    Picture 4: Injection valve ports, standard configuration. Connections:

      1 - From pump

      2 - To metering device

      3 - Plug

      4 - To waste

      5 - From needle seat

      6 - To analytical column


    Keep the vialsampler in the "Needle Up" position.
  5. Close the purge valve and take note of the pressure. Open the purge valve to release the pressure and close the purge valve again for another pressure pulse. Repeat to try and flush out the blockage. Make sure to have Kimwipes or paper towels to collect the mobile phase flow at the needle seat.
  6. Once there is flow backwards through the seat capillary, open the purge valve.
  7. Reinstall the needle seat capillary into port 5, return the capillary that goes to the column to its original position on the injection valve (port 6), flip back the wash port and reinstall the safety cover.
  8. In Lab Advisor under "Injector Steps" click "Needle Into Seat" and then "Reset" to finish the procedure (Picture 5).
    Picture 5: Lab Advisor Injector Steps - Needle Into Seat + Reset.
  9. End the connection from Lab Advisor, and close Lab Advisor.
  10. Open your CDS software and run the HPLC with normal flow rates. Make sure to work in mainpass mode.

 

If the pressure is still high after the backflush, you will need a new seat. Contact Agilent Technologies Support for more help.

 

 

 

Learn more on how to effectively operate and troubleshoot your Agilent InfinityLab Vialsampler:

Vialsamplers Agilent InfinityLab LC Series User Manual

 

 

Keywords: Liquid Chromatography, HPLC, UHPLC, 1260 Infinity II, needle seat backflush, back flush, LC high pressure, overpressure, high backpressure, high back pressure, injectors

  • uhplc
  • LC
  • HPLC
  • injectors
  • 1260 infinity ii
  • sample introduction
  • agilent support insights
  • Share
  • History
  • More
  • Cancel
Anonymous
  • Carlos Vargas
    Carlos Vargas 2 months ago in reply to camelia

    Hello camelia, you can install the basic version of Lab Advisor following the instructions in this article: How to Install Lab Advisor | Agilent Community

    • Cancel
    • Up 0 Down
    • Reply
    • More
    • Cancel
  • camelia
    camelia 2 months ago in reply to Carlos Vargas

    I can not install Lab Advisor without going through a series of approvals and validation hoops. So a document describing how to perform the backflush independently of Lab Advisor is required

    Regards

    • Cancel
    • Up 0 Down
    • Reply
    • More
    • Cancel
  • camelia
    camelia 2 months ago

    It would be useful to have such a document detailed on how to flush the needle seat without the use of labadvisor. Not all customers have labadvisor.