Wiki Troubleshooting High Pressure Issues in Agilent LC Systems
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
Liquid Chromatography requires membership for participation - click to join
  • Wiki
  • About LC and CE
  • Backflushing the needle seat of the Agilent Vialsampler
  • Best Practices for Using an Agilent LC System
  • Can the 1120 and 1220 Series LC Systems be Controlled by the Instant Pilot?
  • Collection of LC (HPLC) Resources
  • Distance to window on CE capillary
  • Forced Cold Start Procedure
  • Fraction Collector Delay Volume
  • General Column Regeneration Procedure
  • How to Add a New System in Lab Advisor
  • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
  • How to change the needle and needle seat in an Agilent G7129x Vialsampler
  • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
  • How to Install Lab Advisor
  • How to perform a cold start for 1200 series LC modules
  • How to Prepare your HPLC for a Long Time Shut Down
  • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
  • How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module
  • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
  • How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
  • How to replace the PTFE frit of the Purge Valve in LC Pumps
  • How To Set Screen Brightness on the Instant Pilot Controller
  • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pumps
  • InfinityLab Quick Connect A-Line Fittings
  • Join the Agilent Biopharma Influencers Community!
  • LC Error Codes
  • LED Status Indicators
  • Method Scouting Wizard Tutorial Videos now available on YouTube!
  • Needle seat backflush procedure
  • Normal Phase seals and wear-in procedure
  • Optical Balance Procedure for Refractive Index Detector
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
  • Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
  • Spectra are not showing for some peaks when collected only with Peak Apex
  • Troubleshooting High Pressure Issues in Agilent LC Systems
  • Troubleshooting peak area variation between injections
  • Use of Multiple Degasser Chambers for Isocratic and Binary pumps without SSV

Troubleshooting High Pressure Issues in Agilent LC Systems

Agilent Support Insights

 

This Information Applies To:  All standard analytical scale Agilent 1100/1200/1260/1290 Infinity / Infinity II LC system configurations:  Single Pump (Isocratic, Quaternary, or Binary) with Manual Purge Valve; no Multisampler G7167; no additional Switching Valves.

 

 

Issue:

When facing high pressure-related problems, like aborted runs, or high pressure error messages, it is important to identify the location of the blockage that is causing high pressure. This information will help you to solve the issue, or to provide the Agilent Service Representative with the necessary information to do so.

Even when part of your whole system configuration is not included in this article, you may still find helpful tips for checking other modules.

 

Steps to follow: 

Be aware that even though it may be unlikely, more than one blockage can occur at the same time. This is especially true for simultaneous blockages of the needle and the needle seat.

Warning: When disconnecting capillaries and starting flow, always have Kimwipe or paper towels ready to collect the mobile phase flow coming out of the open end of the flow path.

Warning: Avoid removing capillaries under high pressure, make sure to wear appropriate protective gear.

 

  1. Check the column:
    1. Disconnect the column outlet and check the pressure (Picture 1). 

      Picture 1: Example of a column compartment (of a G1316 module):

          I - Heat exchanger inlet
      II - column inlet
      III - column outlet
      A - from autosampler/injector
      B - to detector.
         
    2. If the pressure drops drastically (more than a few bar due to simple reduction of capillary backpressure), then the blockage is after the column: Go to step 6, and if it does not point to a solution, go through steps 2 to 5 also.
    3. If the pressure remains high, disconnect the inlet to the column.
    4. If the pressure is fine now (does not rise above the expected values) and you get a flow of solvent out the end of the capillary, the column is clogged. You may try to backflush the column or replace the column as required.
    5. If the pressure remains high, the cause for high pressure is somewhere before the column. Go on to step 2.


  2. Check the method parameters:
    1. Ensure that the upper and lower pressure limits in the method are set to suitable values for the analysis and the system (be aware of the upper pressure limit for your system). In both ChemStation C.01.X and OpenLab 2.X in the system diagram, right-click the pump icon in the dashboard and select Method (see Picture 2).

      Picture 2: System Diagram - Pump module.
    2. Next, insert suitable values for the pressure limits according to your system (compare with Picture 3).


      Picture 3: Pressure Limits - Method parameters.


  3. Check the Pump / Purge Valve:

     

    1. If the pressure limits are correct, open the purge valve and check the pressure (for a flow of 1 mL/min the pressure should be roughly below 2 bar under normal conditions).
    2. If the pressure drops when opening the purge valve, the pump is not clogged (bear in mind that this article only applies to pumps with manual purge valve). Close the purge valve and go on to step 4.
    3. If the pressure is still high, the problem is most likely due to a clogged PTFE frit that resides in the purge valve (see Picture 4). Replace the PTFE frit.

       

      Picture 4: Purge Valve Assembly. 1 - Valve body, 2 - PTFE frit with groove, 3 - Seal cap.
    4. If the preceding step does not solve the issue, remove the purge valve and start the pump. If the pressure does not rise, the purge valve needs to be replaced. If the pressure is still high, the blockage is either in the pump head or between the pump head (or heads) and the purge valve.


  4. Check the Autosampler:
    1. Go to the sampler icon in your software. Switch the injection valve to bypass by right-clicking the sampler in the system diagram and selecting "Switch Valve to Bypass" and start the flow.

    2. If in bypass mode the expected pressure values are obtained, the blockage is in the autosampler flow path (needle, needle seat capillary, sample loop, metering head, rotor seal, or associated capillaries). Go on with step 4.d.

    3. If the pressure is still high in bypass mode, first check the capillary that goes from the pump (purge valve outlet) to the autosampler.
      1. First, remove the capillary connected to the purge valve outlet (Picture 5) and - with the purge valve closed - see if the pressure still rises when starting flow.

        Picture 5: Purge valve outlet capillary.



      2. If the pressure does not rise, and solvent comes out of the purge valve outlet, tighten the connection from Picture 5 again and loosen the other end of the capillary on the injection valve (Picture 6). Start the flow.
      3. If solvent comes out of the capillary's end, the capillary is fine. The blockage is either in the rotor seal or somewhere after the autosampler in the flow path. Reconnect it and go on to step 4.h. 

    4. Disconnect the needle seat capillary tubing from its port of the switching valve (typically port 5, compare with Picture 6) and switch the valve back to main pass.

      Picture 6: Injection valve ports, standard configuration. Connections:

         

      1 - from pump
      2 - to metering device + needle
      3 - plug
      4 - to waste
      5 - from needle seat
      6 - to analytical column.

         

    5. If high pressure is observed again, then there is a blockage before the entry port of the needle seat (typically port 5).
    6. If the pressure is normal and liquid drops out of the capillary, the rotor seal is clogged between port 5 and 6.
    7. Re-attach the needle seat capillary back into port 5.

      Helpful Tip: If the blockage is found to be in the needle seat assembly, sometimes it can be recovered by backflushing the needle seat.

    8. Now remove the capillary going from the injection valve to the column or column compartment by removing the capillary fitting at the injection valve port (See Picture 7).

      Picture 7: Injection valve ports, standard configuration.

    9. If the pressure is normal, the blockage is located after the injection valve (capillary, column compartment, column, detector or associated capillaries).
    10. If high pressure is observed, the blockage is in the rotor seal within the injection valve. Replace the rotor seal.



  5. Check the Thermostatted Column Compartment / Multicolumn Thermostat:
    1. When the pressure in bypass is still high, disconnect the inlet to the column compartment / heat exchanger (Picture 1, Picture 8) and check the pressure. If you have not done step 1 already, do it first.

      Picture 8: Example of a Multicolumn Thermostat (of a G7116 module). Inlet and outlet of the heat exchanger
      (blue) are marked in red.


       

    2. If the blockage is not in the heat exchanger, the capillary or the injection valve are blocked.
    3. Disconnect the end of the capillary at port 6 of the injection valve (see Picture 7) to check if the capillary or the rotor seal is causing the blockage.



  6. If a blockage lies after the column: 
    1. If disconnecting the column outlet releases the high pressure, look for a blockage in the system beyond the column. Check the capillary to the detector, the flow cell, and the waste tubing.
    2. If you cannot locate the source of high pressure after having been through steps 1 to 6, contact your local Agilent Customer Technologies Support representative for further technical support.

 

Learn more on how to effectively troubleshoot your HPLC high pressure issues:

Agilent 1260 Infinity LC Systems with OpenLab CDS ChemStation Edition Learning Path on Agilent University

  • troubleshooting
  • LC
  • Detectors
  • Pumps
  • sample introduction
  • agilent support insights
  • Vacuum Degassers