Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography
  • Technical Areas
  • More
Liquid Chromatography
Wiki How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module
  • Announcements
  • Forum
  • Files
  • Videos
  • Wiki
  • More
  • Cancel
  • New
Liquid Chromatography requires membership for participation - click to join
  • Wiki
  • About LC and CE
  • Backflushing the needle seat of the Agilent Vialsampler
  • Best Practices for Using an Agilent LC System
  • Can the 1120 and 1220 Series LC Systems be Controlled by the Instant Pilot?
  • Collection of LC (HPLC) Resources
  • Distance to window on CE capillary
  • Forced Cold Start Procedure
  • Fraction Collector Delay Volume
  • General Column Regeneration Procedure
  • How to Add a New System in Lab Advisor
  • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
  • How to change the needle and needle seat in an Agilent G7129x Vialsampler
  • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
  • How to Install Lab Advisor
  • How to perform a cold start for 1200 series LC modules
  • How to Prepare your HPLC for a Long Time Shut Down
  • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
  • How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module
  • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
  • How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
  • How to replace the PTFE frit of the Purge Valve in LC Pumps
  • How To Set Screen Brightness on the Instant Pilot Controller
  • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pumps
  • InfinityLab Quick Connect A-Line Fittings
  • Join the Agilent Biopharma Influencers Community!
  • LC Error Codes
  • LED Status Indicators
  • Method Scouting Wizard Tutorial Videos now available on YouTube!
  • Needle seat backflush procedure
  • Normal Phase seals and wear-in procedure
  • Optical Balance Procedure for Refractive Index Detector
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
  • Removing and re-installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD Modules
  • Spectra are not showing for some peaks when collected only with Peak Apex
  • Troubleshooting High Pressure Issues in Agilent LC Systems
  • Troubleshooting peak area variation between injections
  • Use of Multiple Degasser Chambers for Isocratic and Binary pumps without SSV

How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module

Agilent Support Insights

 

This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules

 

Issue:

Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module.

 

Steps To Follow

When loosening the pump head blots, make sure that the bolts are loosening evenly left and right. Do not loosen one pump bolt fully while leaving the other bolt tight as it may cause misalignment of the pump head and snapped pistons.

  1. Remove the tubing from your purge valve to your autosampler.



  2. Power off the pump and remove the waste tubing from the purge valve.
  3. If you have an Agilent 1260 Infinity Isocratic or Quaternary Pump, remove the tubing to your passive inlet valve (PIV). If you have a Agilent 1260 Infinity Binary Pump, remove the tubing to your Active Inlet Valve (AIV).
  4. Remove the tubing from your outlet ball valve on the top of the pump head to your damper.
  5. Next remove the tubing from your damper to piston 2, which is located on the bottom of the pump head.


  6. To separate the pump head from the metering drive, use the provided Allen wrench to loosen the pump head bolts 2 - 3 turns each, moving side to side, to make sure you are loosening the bolts evenly (see next picture). Only loosening one side can twist the pump head on the metering drive and can lead to piston breakage.
  7. Once the bolts are removed, slide the pump head gently off the metering drive. Be careful to keep the back of the pump head tilted upwards to prevent the pistons falling out of the pump head and breaking.

Do not remove the small bead of grease you see on the face of each piston at the back of the pump head. This is required for lubrication of the piston face against the metering drive.


 

Learn more on how to effectively troubleshoot your Agilent 1260 Infinity Pump:

HPLC-1260-2230z - Changing Pump Seals in Agilent 1260 Binary and Quaternary and Isocratic Pumps e-Learning course at Agilent University

 

 

Keywords: Agilent 1260 Infinity Pump, pump head removal, pump head maintenance, 1260, Binary, Quaternary, Isocratic, lc, 1260 infinity, hplc, pumps

 

 

  • g1312b
  • LC
  • g1311b
  • g7112b
  • g1311c
  • g1310b
  • Pumps
  • 1260 infinity
  • g7111b
  • 1260 infinity ii
  • g7110b
  • agilent support insights
  • g7111a
  • Share