Needle wash solvent in needle?

I am new user on agilent 1290 system running on labsolution software. I run gradient of water methanol 10-99%B on C18. The method works fine when i dont use the needle wash but if i use needle wash in wash port and wash solvent 100% meoh then there is no retention on column my peak is in void volume instead of the normal 5min. My understanding was that needle wash just wash outside of needle before injection. It seems that a large amount of wash solvent is injected with the sample but I dont hunderstand how that can happen. There are no options in the software that suggest the wash solvent is drawn into syringe.

 Can anyone explain whats going on and how my wash solvent can be injected with my sample?

There has to be a large amount of meoh since it is not just peak fronting but actually entire signal is in void volume.

Using less strong wash solvent would be possible but would still like to understand whats happening.

 

Parents Reply Children
  • Thanks for reply. Will have to check the module part nr but its infinity model similar to G4226-90001. I cant see that its equilibration problem since method works great when I dont use needle wash or use 20%meoh as needle wash.

    I did find a tech note that this wash design was not suitable for strong wash solvents since some solvent on needle exterior is injected.  I injected 2ul sample in water on 20%meoh mobilephase so perhaps a small volume of meho on needle exterior was the problem. The column was 2.1"100mm WatersT3, a c18 column. On the systems i used before i could use strong needle wash solvents but now i will consider that it can influence lc method.

    The needle external wash was setup to use flushport for 6 seconds

    On request i will provide the link to the evaluation of needle wash station design title

    "NEEDLE WASH MECHANISM AND ITS
    IMPACT ON HILIC ANALYSIS"

  • Hi  

    You could test your wash solvent mixing hypothesis by dissolving your sample in methanol or methanol/water, assuming that the analyte is soluble in methanol; I assume it must be, since you're using methanol as a wash solvent.  If you see the poor retention with the organic injection solvent then you have an answer.

    How does that sound?

    /Andy

  • Thanks thats a good suggestion. The analyte is caffeine for a system commissioning test. The system works fine and no carry over problem so very strong wash solvents not needed. Still good to understand how system works.

Was this helpful?