Needle wash solvent in needle?

I am new user on agilent 1290 system running on labsolution software. I run gradient of water methanol 10-99%B on C18. The method works fine when i dont use the needle wash but if i use needle wash in wash port and wash solvent 100% meoh then there is no retention on column my peak is in void volume instead of the normal 5min. My understanding was that needle wash just wash outside of needle before injection. It seems that a large amount of wash solvent is injected with the sample but I dont hunderstand how that can happen. There are no options in the software that suggest the wash solvent is drawn into syringe.

 Can anyone explain whats going on and how my wash solvent can be injected with my sample?

There has to be a large amount of meoh since it is not just peak fronting but actually entire signal is in void volume.

Using less strong wash solvent would be possible but would still like to understand whats happening.

 

Was this helpful?