I have changed the cassette, changed the capillary, changed every part of it, looked at the connections, restarted my computer and turned off the CE many times.
Find maintenance, best practice, and troubleshooting articles.
I have changed the cassette, changed the capillary, changed every part of it, looked at the connections, restarted my computer and turned off the CE many times.
Hi anoopsidhu10 welcome to the Community!
I moved your question to the LC Forum for better visibility.
Cheers,
Kristen
what happens if you insert the cassette without a capillary installed and only the green interface installed. . Make sure the alignment interface snaps onto the ceramic interface of the detector. Check the lamp life and intensity.
checked the lamp life and intensity, it says "intensity at test" is zero, do you know how I can fix the intensity?
did you do this without the capillary? how many hours does the lamp have? Have you replaced the lamp? If yes, then you may need to call your local technical support for possible on-site repair.