I can connect our Agilent 1290 system to Lab Advisor
I cannot connect the Agilent 1290 ELSD to it though.
Both have different IP addresses.
Ideas anyone?
Find maintenance, best practice, and troubleshooting articles.
I can connect our Agilent 1290 system to Lab Advisor
I cannot connect the Agilent 1290 ELSD to it though.
Both have different IP addresses.
Ideas anyone?
Hello,
I am going to move this post to the LC forum since that is the group that supports the Lab Advisor software.
Marty
Many thanks
Hello monro -
The 1290 ELSD will communicate with Lab Advisor only if you install the ELSD Add-0n. Within Lab Advisor click on Configuration, then click into the Market, look for Agilent ELSD Lab Advisor addon and install it. You may need to Update Lab advisor for compatibility.
Next, go back to System Overview and use the Add System button. change the Instrument Type to ELSD. Enter the correct address for communication. Be aware the only function available for the ELSD is Firmware update. There is no ability to control or perform diagnostics on this detector with Lab Advisor.
-Howard