Does anyone know a good way to clear this error or trouble shoot this error?
The error means that the maximum stroke length of your pump drive is being shorter than it should be.
This can be either due to mechanical issues like a broken piston or an offset in the pump drive gear leading to a physical block before the normal amount of motor steps can be reached.
Another root cause may be a blockage in the liquid path before the pressure sensor.
For troubleshooting this you need to open all valves and flow connections between the pump heads and pressure sensor before powering on your pump.
If the pump is able to initialize you need to check for the blocked part by re-connecting the parts step by step in a logical way to see when it starts getting blocked again.
I tried opening the capillary connections. Seems there is no blocks. In g7120A, I set 100% flow in A1 or A2 and B1 or B2, but the same error is coming,"pump drive stroke blocked". I cannot be sure of which drive is the problematic one. Can anyone help?
I tried opening the capillary connections. Seems there is no blocks. In g7120A, I set 100% flow in A1 or A2 and B1 or B2, but the same error is coming,"pump drive stroke blocked". I cannot be sure of which drive is the problematic one. Can anyone help?