I had a capillary break off of the 67116-60015 quick connect heat exchanger in my 1290 column compartment. How do I replace it? A wrench won't fit.
Find maintenance, best practice, and troubleshooting articles.
I had a capillary break off of the 67116-60015 quick connect heat exchanger in my 1290 column compartment. How do I replace it? A wrench won't fit.
Hi
I will suggest to call your local agilent support..They will come with their dedicated tools and can fix it easily...
You should be able to use the fitting mounting tool 5043-0915 to take out the G7116-60015 from the heating block.
It should have come in the accessory kit of the instrument at installation. The tool is only a few centimeters long. see below
Then you can use 2 quarter inch wrenches to remove and replace the capillary.