Hi, I am trying to isolate the source of contamination. Is there a way to set up a "no injection" gradient run? I am using a 1290 Infinity II, 6546 QTOF


Parents Reply
  • Hi

    What SW are you using? You should be able to set up a wash method and include it in your worklist/sequence table. However for it to automatically not pump solvent through the column, you will need a column switching valve set up in your column compartment (and set the valve position in the wash method accordingly). Without a column switching valve, you will need to manually swap the column out with union.

    Would you be able to share the model and part # of your column? It is best to check this with our technical experts in the Consumables area.