Hello,
i need to additional installing the low pressure mixer (G4204-81090) at the quaternary pump G71117B - to improve the perfomace.
The Pump has a AIV - ist there possible?
Is it possible to change the AIV to PIV?
thanks,
Hello,
i need to additional installing the low pressure mixer (G4204-81090) at the quaternary pump G71117B - to improve the perfomace.
The Pump has a AIV - ist there possible?
Is it possible to change the AIV to PIV?
thanks,
Hi treavour
You can certainly change AIV to PIV, but I'm not sure that it us possible to install the G4204-81090 "inlet weaver" into a 1260 quaternary pump. It's probably better to install additional mixing after the pump head, rather than before it.
Andy