I have a issue of high pressure for 1290 Infinity II flexible pump and notice leak from high pressure housing filter. How to approach to resolve this issue?
Find maintenance, best practice, and troubleshooting articles.
I have a issue of high pressure for 1290 Infinity II flexible pump and notice leak from high pressure housing filter. How to approach to resolve this issue?
Hi
Check above link to resolve the issue..
The filter inside the high pressure filter housing is probably clogged and needs to be replaced.
See page 217: https://www.agilent.com/cs/library/usermanuals/public/G7104-G7131-1290-Flexible-Pumps-UseMa-en-SD-29000017.pdf
the part number for the filter frit is: 5067-5716 2/pk
Don't forget that you can keep these filters in good condition by using the "Flush Filter On" action from the popup menu on the pump.
Andy
Only the inline filter can be automatically backflushed. If it is not present, please do not use the flush filter on function because this could damage the valve. It seems that mkaila is referring to the outlet filter because this is also called the high pressure outlet filter. This filter cannot be automatically backflushed. But it is true for users with inline filters, the flush filter on action is useful for keeping these filters in good condition.