I'm using a Waters Gen-Pak column using 1M NaCl salt buffer as a mobile phase. The max recommended pressure for this kind of column is 4000psi. What kind of pressure should I expect during a run?
-Thank you
Find maintenance, best practice, and troubleshooting articles.
Hello Everyone,
I'm using a Waters Gen-Pak column using 1M NaCl salt buffer as a mobile phase. The max recommended pressure for this kind of column is 4000psi. What kind of pressure should I expect during a run?
-Thank you
Pressure depends on your HPLC capillaries (diameter, length), on the particle size of the column and the flow of the mobile phase. There is no exact numbers we can give you. All we can really say is to monitor day from day 1 and see if it goes up or down.