Hello Everyone,
After using the 1260 infinity system for purification what solvent can I use to clean and store the system in for short and long term without use.
-Thank you.
Find maintenance, best practice, and troubleshooting articles.
Hello Everyone,
After using the 1260 infinity system for purification what solvent can I use to clean and store the system in for short and long term without use.
-Thank you.
Normally we suggest storing in IPA or MeOH.
Here is a good guide for most tasks with HPLC
www.agilent.com/.../BestPractice_en.pdf
Where it says to store in recommended solvent, I would suggest IPA
Normally we suggest storing in IPA or MeOH.
Here is a good guide for most tasks with HPLC
www.agilent.com/.../BestPractice_en.pdf
Where it says to store in recommended solvent, I would suggest IPA