DAD flow cell

I have a leaking DAD flow cell in my 1200 series HPLC, is there an option for repair...or does it need to be replaced?

Parents Reply Children
  • You may want to order the rebuild kit pn  G1315-68712

    Because of high pressure or use the widows might crack and leak.  The rebuild kit comes with spare gaskets as well.

    Does the outlet of the flow cell go to waste?  or another module- detector or fraction collector?  We also have pressure relief valves to protect the flow cell G4212-68001

Was this helpful?