I have a leaking DAD flow cell in my 1200 series HPLC, is there an option for repair...or does it need to be replaced?
Find maintenance, best practice, and troubleshooting articles.
I have a leaking DAD flow cell in my 1200 series HPLC, is there an option for repair...or does it need to be replaced?
I moved this post to the Liquid Chromatography group where it will get more visibility.
There is a part number to rebuild the flow cells for some of the DADs- the G1315X and G7115X model DAD
What is the part number for the flow cell you have...? it is written on the flow cell.
Here is the Infinity Lab LC supplies catalog https://www.agilent.com/cs/library/catalogs/public/InfinityLab_5991-8031EN_Catalog_LR.pdf
See pages 76-78 -it has the flow cell rebuild kits
I have G1315-60022.
It's leaking from the middle of the cell, maybe indicates quartz needs replacement? or could it be another part. just a gasket?
thanks so much!
You may want to order the rebuild kit pn G1315-68712
Because of high pressure or use the widows might crack and leak. The rebuild kit comes with spare gaskets as well.
Does the outlet of the flow cell go to waste? or another module- detector or fraction collector? We also have pressure relief valves to protect the flow cell G4212-68001