Has anyone seen the error code 506 on a 1260 infinity II LC? Please see the pictures below for further information.
Over many years and lots of experience with Agilent LC we have never experienced this before. Kind regards, Stefan.
Find maintenance, best practice, and troubleshooting articles.
Has anyone seen the error code 506 on a 1260 infinity II LC? Please see the pictures below for further information.
Over many years and lots of experience with Agilent LC we have never experienced this before. Kind regards, Stefan.
Hey all,
This is what occurs when the firmware revision of the instant pilot is too old. Please update the firmware revision of the instant pilot (G4208A) to B.07.23 and this should solve the problem.