In our Infinity 1260 system, the robotic arm getting struck during sampling, especially in last few lines. The issue started on its own and the attached video shows the issue.
Find maintenance, best practice, and troubleshooting articles.
In our Infinity 1260 system, the robotic arm getting struck during sampling, especially in last few lines. The issue started on its own and the attached video shows the issue.
Did anyone do any maintenance of transport assy of autosampler or not.
2nd you need lab advisor to run alignment of trsnsport assy..
3rd when you changed finger cap of robotic arm..sometimes these green finger cap of robotic arm give problem..during PM, basically, this finger cap should be changed.
After this , problem of motor etc. can be checked but for that you have to contact agilent to get service..
No maintenance done nor anything changed..., but the system was ideal for couple of months.
Thanks for your response.., and today Agilent engineer attending to it.
Thanks for the response !
No maintenance done nor anything changed..., but the system was ideal for couple of months.
Thanks for your response.., and today Agilent engineer attending to it.
Thanks for the response !