Hi,today I have problem with bubbles.
I can't eliminate them,I change the flasks,I used ultrasonic bath.
No way.
Hi,today I have problem with bubbles.
I can't eliminate them,I change the flasks,I used ultrasonic bath.
No way.
You experienced a phenomenon called "cavitation". This is due to dirty or clogged solvent filter (picture). For assurance you can TEMPORARY remove solvent filter and run the pump. If the bubbles disappear…
Do not try to clean the glass inlet frits, e.g. by sonicating them or boiling in Nitric Acid. Just bin them and go for new ones. Alternatively you can order SST inlet frits, which can be cleaned and re…
Thank you.
Do not try to clean the glass inlet frits, e.g. by sonicating them or boiling in Nitric Acid. Just bin them and go for new ones. Alternatively you can order SST inlet frits, which can be cleaned and re-used. Poresize is 20um vs. ~14um.
Thank you.
Thank you.