Hi all, How can I stop my LC after finished a sequence. I'm running a gradient and want to add a stop time.
In chemstation, this is under sequence parameters and then there is a location titled post-sequence macro. In openlab 2.3 or higher, you can choose a shutdown method in the run queue, see 1:40 OpenLab CDS for ChemStation users - Submitting and managing runs - YouTube. In older versions of openlab 2, there are shutdown parameters in the openlab control panel and there is always the possibility to create an actual method shutting down the instrument (besides lamp, lamp shutdown is controlled in the control panel).
In chemstation, this is under sequence parameters and then there is a location titled post-sequence macro. In openlab 2.3 or higher, you can choose a shutdown method in the run queue, see 1:40 OpenLab CDS for ChemStation users - Submitting and managing runs - YouTube. In older versions of openlab 2, there are shutdown parameters in the openlab control panel and there is always the possibility to create an actual method shutting down the instrument (besides lamp, lamp shutdown is controlled in the control panel).